How to build a new stylus system

Build and qualify a new stylus system in ZEISS CALYPSO for fixed star, RDS, and XTR heads.

⭐

Before you start: Qualify the reference sphere position with the Master Probe. Redo this whenever the sphere is moved or the CMM has been powered down.

Before you build any stylus system:

  • Finger tight is plenty. Don’t over-tighten.
  • Name it right the first time. Use a short, descriptive, consistent name (for example, 2.0x30+40 for a 2 mm x 30 mm stylus with a 40 mm extension). You can’t rename a stylus system later. See Stylus System Naming Conventions.
  • Named one wrong? You can’t rename a stylus system, but you can copy it under a new name. In Stylus system management, select it, click Copy, and enter the new name. The copy stays qualified, but any program that uses the old one needs to be reassigned to the copy.
  • Use the standard stylus numbers #1–#5, even if a direction is missing. Use #6 and up for any other directions.
#
Direction
RDS angle
XTR angle
1
-Z
A0 B0
A0
2
+Y
A90 B-90
A-90
3
+X
A0 B90
A0
4
-Y
A90 B90
A90
5
-X
A0 B-90
A180

Pick your probe head below to get started.

Fixed Star Probe

Notion image
Notion image
  • VAST XT
  • Duramax
  • O-Inspect
  1. Assemble the stylus system.
  1. Open the Stylus System window.
  1. Click Manual Stylus System Change and load the stylus system into the probe head. In the Select stylus system window, click New.
  1. Enter the stylus system name in the first box.
    1. Notion image
  1. Enter a name for the first stylus and set its stylus number from the table above.
    1. Notion image
  1. Click New stylus to add each additional stylus.
    1. Notion image
  1. Select a stylus from the Stylus name list and click Qualify Stylus to qualify or “calibrate” it. Take the first manual point toward the center of the reference sphere, in the direction of the stylus shaft. Repeat for each stylus.
    1. Notion image
      Notion image

RDS Rotating Probe

Notion image

Which mode should you use?

  • CAA mode: Fastest setup. Best for parts that need many probe angles, such as complex castings or large aerospace parts.
  • Non-CAA mode: More accurate. Best for tight tolerances.

Check if your probe uses CAA mode

In the Stylus System window, Mode shows RC-CAA. There is usually only one stylus.

Notion image
Notion image

CAA Mode

  1. First-time setup only: load the Master Probe and run Fitting Position from the Stylus System window. You only need to do this once per RDS installation.
  1. Assemble the stylus system.
  1. Open the Stylus System window.
  1. Click Manual Stylus System Change and load the stylus system. In the Select stylus system window, click New.
    1. Notion image
  1. Enter the stylus system name in the first box.
  1. Enter the same name for the stylus.
  1. Select the RDSCAA checkbox, then click OK.
  1. Click Qualify Stylus to qualify or “calibrate” it.

CALYPSO physically qualifies 12 positions and calculates all other angles. You don’t need to add individual angles.

Notion image

Assign an angle to a feature: Go to Measurement Plan Editor > Stylus System > RDS Angles. This menu appears only when a CAA stylus system is loaded. Select the feature(s), then pick an angle from the list. For any other angle, select 0.0° / 0.0°, edit the A/B angles, and press Enter.

💡

The RDS rotates in 2.5° increments.

Non-CAA Mode

Here, each stylus is really one RDS angle.

  1. Assemble the stylus system.
  1. Open the Stylus System window.
  1. Click Manual Stylus System Change and load the stylus system. In the Select stylus system window, click New.
    1. Notion image
  1. Enter the stylus system name in the first box.
  1. Name the first stylus -Z A0 B0 (#1).
  1. If the head isn’t at A0 B0, click Rotate Stylus, enter A0 B0, then click Rotate the axes on the machine. Make sure the path is clear!
  1. Click Qualify Stylus to qualify or “calibrate” it.
  1. Click New stylus and name it +Y A90 B-90 (#2).
    1. Notion image
  1. Click Rotate Stylus, enter A90 B-90, then click Rotate the axes on the machine.
    1. Notion image
  1. Click Qualify Stylus.
  1. Repeat steps 8–10 for #3–#5, then add any other angles (#6 and up, in any order).
⏱

Faster for many angles: List Qualification. Click Rotate Stylus > Styli List tab and enter all stylus names and A/B angles in one table (or copy the list from an existing RDS stylus system). Then, in the Probing System Qualification window, click CNC Probing System Qualification > List Qualification. You take one manual point, and CALYPSO qualifies every angle automatically. One catch: skip this if small styli need a custom Probing Dynamic.

A/B angle reference:

Notion image

XTR Rotating Probe

Build a star probe with styli in the +X and -Z directions. Skip the -Z stylus if you don’t need it. The XTR rotates the +X stylus to reach +Y, -Y, and -X.

  1. Assemble the stylus system.
  1. Open the Stylus System window.
  1. Click Manual Stylus System Change and load the stylus system. In the Select stylus system window, click New.
    1. Notion image
  1. Enter the stylus system name in the first box.
  1. Name the first stylus and set its stylus number from the table above.
    1. Notion image
  1. Click Qualify Stylus to qualify or “calibrate” the first stylus.
  1. Click New stylus to add #2 (+Y).
    1. Notion image
  1. Click Rotate Stylus, rotate to A-90, then click Rotate the axes on the machine. Make sure the path is clear!
    1. Notion image
      Notion image
      Notion image
  1. Click Qualify Stylus to qualify the +Y (#2) stylus.
  1. Repeat steps 7–9 for the remaining directions in the table above.

After you qualify

  • Check the results. A quick look at the sigma (S) value after each qualification catches problems early. If it’s higher than normal, clean the stylus and sphere, then requalify.
  • Using a probe rack? Assign the stylus system to its holder: CNC Stylus System Change > select the holder > Changer > Set stylus system to stylus system holder.